Page 1 of 1
VCP1001 Resolution
Posted: Wed Jan 24, 2024 10:41 am
by ProtonTim
The VCP1001_0 says 10bits on the case, but in the specs it says it has 200uV resolution across a +5/-5 range. 10V/2^10 = 9.7mV, what am I missing?
Re: VCP1001 Resolution
Posted: Thu Jan 25, 2024 9:40 am
by mparadis
The VCP1001 is oversampled, so the effective resolution is much higher than the base 10-bit resolution of the ADC. We recognized that this was confusing so we removed it from the product name, but the enclosure couldn't be changed for our existing stock.
Re: VCP1001 Resolution
Posted: Thu Jan 25, 2024 2:35 pm
by ProtonTim
Thank you for the clarification.
The 20 bit ADC (VCP1000_0) states that is has a Measurement Error Max 0.01%. Does that mean the reported measurements can be off by as much as 0.0001 * 80V = 8mV even though it has 77μV resolution?
Would this error only occur at this magnitude at full scale?
If I were to use the VCP1000_0 to measure two voltages that were offset by 4.000V what is the worst differential error I should expect with such a measurement? (0.4mV?)
Thank you for your help on this.