Combine Phidget 1046 with 1002

Post by engFM »

Hello everyone

For a test setup I need to measure the applied force continuously. This is done by a load cell from Honeywell Series 53 5000lb.

I would like to read this load cell with the phidget 1046 but this has only +-5 V output voltage whereas the load cell requires +-10V.

My idea was that I could supply the sensor with the phidget 1002 as this would deliver the required voltage and current.

Is it possible to do this?
And how would the circuit look like?

Thank you for your help,
Marcel
User avatar
mparadis
Site Admin
Posts: 660
Joined: Fri Oct 28, 2011 12:17 pm
Contact:

Re: Combine Phidget 1046 with 1002

Post by mparadis »

Keep in mind that the 1046 is only designed to read up to a maximum of about 4.75V, so even if you do power a load cell with 10V, the voltage reading coming out of the cell might be too high.