Temperature Sensor delay after re-attaching
Posted: Thu Jul 20, 2017 2:42 am
I have the 1048 and the 1051 TemperatureSensors but I noticed something a little strange. I am able to read from the temperature sensors without a problem.
However, when I physically detach the sensors and then reattach. The reading taken immediately after the attached event produces very low numbers roughly between 0-2 degrees. There is no error being returned from the call.
I also have the IR TemperatureSensor and it doesn't have this same problem.
Retrieving the correct temperature takes approximately 500ms. After that everything is back to normal. Is this delay with getting a correct reading after detaching and re-attaching the correct behaviour?